Fluorescence interference contrast (FLIC) microscopy is a
microscopicMicroscopy is the technical field of using microscopes to view samples or objects. There are three well-known branches of microscopy, optical, electron and scanning probe microscopy....
technique developed to achieve z-resolution on the nanometer scale.
FLIC occurs whenever fluorescent objects are in the vicinity of a reflecting surface (e.g. Si wafer). The resulting interference between the direct and the reflected light leads to a double sin
2 modulation of the intensity, I, of a fluorescent object as a function of distance, h, above the reflecting surface. This allows for the
nanometer height measurements.
FLIC microscope is well suited to measuring the topography of a membrane that contains fluorescent
probes e.g. an artificial
lipid bilayerA lipid bilayer is a thin membrane made of two layers of lipid molecules. These membranes are flat sheets that form a continuous barrier around cells. The cell membrane of almost all living organisms and many viruses are made of a lipid bilayer, as are the membranes surrounding the cell nucleus and...
, or a living
cell membraneThe cell membrane is the biological membrane separating the interior of a cell from the outside environment....
or the structure of fluorescently labeled
proteinProteins are organic compounds made of amino acids arranged in a linear chain and folded into a globular form. The amino acids in a polymer chain are joined together by the peptide bonds between the carboxyl and amino groups of adjacent amino acid residues...
s on a surface.
General two layer system
The optical theory underlying FLIC was developed by Armin Lambacher and Peter Fromherz. They derived a relationship between the observed fluorescence intensity and the distance of the fluorophore from a reflective
siliconSilicon is the most common metalloid. It is a chemical element, which has the symbol Si and atomic number 14. A tetravalent metalloid, silicon is less reactive than its chemical analog carbon...
surface.
The observed fluorescence intensity, , is the product of the excitation probability per unit time, , and the probability of measuring an emitted photon per unit time, . Both probabilities are a function of the fluorophore height above the silicon surface, so the observed intensity will also be a function of the fluorophore height. The simplest arrangement to consider is a fluorophore embedded in silicon dioxide (refractive index ) a distance
d from an interface with silicon (refractive index ). The fluorophore is excited by light of wavelength and emits light of wavelength . The unit vector gives the orientation of the transition
dipoleIn physics, there are two kinds of dipoles:*An electric dipole is a separation of positive and negative charges. The simplest example of this is a pair of electric charges of equal magnitude but opposite sign, separated by some, usually small, distance. A permanent electric dipole is called an...
of excitation of the fluorophore. is proportional to the squared projection of the local
electric fieldIn physics, the space surrounding an electric charge or in the presence of a time-varying magnetic field has a property called an electric field. This electric field exerts a force on other electrically charged objects...
, , which includes the effects of
interferenceIn physics, interference is the addition of two or more waves that results in a new wave pattern.Interference usually refers to the interaction of waves that are correlated or coherent with each other, either because they come from the same source or because they have the same or nearly the same...
, on the direction of the transition dipole.
The local electric field, , at the fluorophore is affected by interference between the direct incident light and the light reflecting off the silicon surface. The interference is quantified by the phase difference given by
is the angle of the incident light with respect to the silicon plane normal. Not only does interference modulate , but the silicon surface does not perfectly reflect the incident light. Fresnel coefficients give the change in amplitude between an incident and reflected wave. The
Fresnel coefficientsIn optics, Fresnel diffraction or near-field diffraction is a process of diffraction that occurs when a wave passes through an aperture and diffracts in the near field, causing any diffraction pattern observed to differ in size and shape, depending on the distance between the aperture and the...
depend on the angles of incidence, and , the indices of refraction of the two mediums and the
polarizationPolarization is a property of waves that describes the orientation of their oscillations. This article primarily covers the polarization of electromagnetic waves such as light, although other types of wave also exhibit polarization....
direction. The angles and can be related by
Snell's LawIn optics and physics, Snell's law , is a formula used to describe the relationship between the angles of incidence and refraction, when referring to light or other waves passing through a boundary between two different isotropic media, such as water and glass...
. The expressions for the reflection coefficients are:
TE refers to the component of the electric field perpendicular to the plane of incidence and TM to the parallel component (The incident plane is defined by the plane normal and the propagation direction of the light). In
cartesianA Cartesian coordinate system specifies each point uniquely in a plane by a pair of numerical coordinates, which are the signed distances from the point to two fixed perpendicular directed lines, measured in the same unit of length....
coordinates, the local electric field is
is the polarization angle of the incident light with respect to the plane of incidence. The orientation of the excitation dipole is a function of its angle to the normal and azimuthal to the plane of incidence.
The above two equations for and can be combined to give the probability of exciting the fluorophore per unit time .
Many of the parameters used above would vary in a normal experiment. The variation in the five following parameters should be included in this theoretical description.
- The coherence
In physics, coherence is a property of waves, that enables stationary interference. More generally, coherence describes all properties of the electronic correlation between physical quantities of a wave....
of the excitation light
- The incident angle of excitation light
- Polarization angle of the excitation light
- The angle of transition dipole of the fluorophore
- The wavelength of the excitation light
The squared projection must be averaged over these quantities to give the probability of excitation . Averaging over the first 4 parameters gives
Normalization factors are not included. is a distribution of the orientation angle of the fluorophore dipoles. The
azimuthalIn mathematics, the polar coordinate system is a two-dimensional coordinate system in which each point on a plane is determined by a distance from a fixed point and an angle from a fixed direction....
angle and the polarization angle are integrated over analytically, so they no longer appear in the above equation. To finally obtain the probability of excitation per unit time, the above equation is integrated over the spread in excitation wavelength, accounting for the intensity and the extinction coefficient of the fluorophore .
The steps to calculate are equivalent to those above in calculating except that the parameter labels
em are replaced with
ex and
in is replaced with
out.
The resulting fluorescence intensity measured is proportional to the product of the excitation probability and emission probability
It is important to note that this theory determines a proportionality relation between the measured fluorescence intensity and the distance of the fluorophore above the reflective surface. The fact that it is not an equality relation will have a significant effect on the experimental procedure.
Experimental Setup
A silicon wafer is typically used as the reflective surface in a FLIC experiment. An
oxideThe chemical compound silicon dioxide, also known as silica , is an oxide of silicon with a chemical formula of ' and has been known for its hardness since antiquity. Silica is most commonly found in nature as sand or quartz, as well as in the cell walls of diatoms...
layer is then thermally grown on top of the silicon wafer to act as a spacer. On top of the oxide is placed the fluorescently labeled specimen, such as a lipid membrane, a cell or membrane bound proteins.
With the sample system built, all that is needed is an epifluorescence microscope and a
CCDA charge-coupled device is a device for the movement of electrical charge, usually from within the device to an area where the charge can be manipulated, for example conversion into a digital value. This is achieved by "shifting" the signals between stages within the device one at a time...
camera to make quantitative intensity measurements.
The silicon dioxide thickness is very important in making accurate FLIC measurements. As mentioned before, the theoretical model describes the
relative fluorescence intensity measured versus the fluorophore height. The fluorophore position cannot be simply read off of a single measured FLIC curve. The basic procedure is to manufacture the oxide layer with at least two known thicknesses (the layer can be made with
photolithographicOptical lithography, is a process used in microfabrication to selectively remove parts of a thin film or the bulk of a substrate. It uses light to transfer a geometric pattern from a photo mask to a light-sensitive chemical photo resist, or simply "resist," on the substrate. A series of chemical...
techniques and the thickness measured by
ellipsometryEllipsometry is a versatile and powerful optical technique for the investigation of the dielectric properties of thin films....
). The thicknesses used depends on the sample being measured. For a sample with fluorophore height in the range of 10 nm, oxide thickness around 50 nm would be best because the FLIC intensity curve is steepest here and would produce the greatest contrast between fluorophore heights. Oxide thickness above a few hundred nanometers could be problematic because the curve begins to get smeared out by polychromatic light and a range of incident angles. A ratio of measured fluorescence intensities at different oxide thicknesses is compared to the predicted ratio to calculate the fluorophore height above the oxide .
The above equation can then be solved numerically to find .
Imperfections of the experiment , such as imperfect reflection, nonnormal incidence of light and polychromatic light tend to smear out the sharp fluorescence curves. The spread in incidence angle can be controlled by the
numerical apertureIn optics, the numerical aperture of an optical system is a dimensionless number that characterizes the range of angles over which the system can accept or emit light...
(N.A.). However, depending on the numerical aperture used, the experiment will yield good lateral
resolutionOptical resolution describes the ability of an imaging system to resolve detail in the object that is being imaged.An imaging system may have many individual components including a lens and recording and display components...
(x-y) or good vertical resolution (z), but not both. A high N.A. (~1.0) gives good lateral resolution which is best if the goal is to determine long range topography. Low N.A. (~0.001), on the other hand, provides accurate z-height measurement to determine the height of a fluorescently labeled molecule in a system.
Analysis
The basic analysis involves
fittingCurve fitting is the process of constructing a curve, or mathematical function, that has the best fit to a series of data points, possibly subject to constraints. Curve fitting can involve either interpolation, where an exact fit to the data is required, or smoothing, in which a "smooth" function...
the intensity data with the theoretical model allowing the distance of the fluorophore above the oxide surface to be a free parameter.
The FLIC curves shift to the left as the distance of the fluorophore above the oxide increases. is usually the parameter of interest, but several other free parameters are often included to optimize the fit. Normally an amplitude factor (a) and a constant additive term for the background (b) are included. The amplitude factor scales the relative model intensity and the constant background shifts the curve up or down to account for fluorescence coming from out of focus areas, such as the top side of a cell. Occasionally the numerical aperture (N.A.) of the microscope is allowed to be a free parameter in the fitting. The other parameters entering the optical theory, such as different indices of refraction, layer thicknesses and light wavelengths, are assumed constant with some uncertainty.
A FLIC chip may be made with oxide terraces of 9 or 16 different heights arranged in blocks. After a fluorescence image is captured, each 9 or 16 terrace block yields a separate FLIC curve that defines a unique . The average is found by compiling all the values into a histogram.
The statistical error in the calculation of comes from two sources: the error in fitting of the optical theory to the data and the uncertainty in the thickness of the oxide layer.
Systematic errorSystematic errors are biases in measurement which lead to the situation where the mean of many separate measurements differs significantly from the actual value of the measured attribute. All measurements are prone to systematic errors, often of several different types...
comes from three sources: the measurement of the oxide thickness (usually by ellipsometer), the fluorescence intensity measurement with the CCD, and the uncertainty in the parameters used in the optical theory. The systematic error has been estimated to be .