All Topics  
Electrosurgery

 

   Email Print
   Bookmark   Link






 

Electrosurgery



 
 
Electrosurgery is the application of a high-frequency electric current to biological tissue as a means to cut, coagulate, desiccate, or fulgurate
Fulguration

Fulguration, also called electrofulguration, is a procedure to destroy tissue using an electric current....
 tissue. (These terms are used in specific ways for this methodology-- see below). Its benefits include the ability to make precise cuts with limited blood loss. Electrosurgical devices are frequently used during surgical operations helping to prevent blood loss in hospital operating rooms or in outpatient procedures.






Discussion
Ask a question about 'Electrosurgery'
Start a new discussion about 'Electrosurgery'
Answer questions from other users
Full Discussion Forum



Encyclopedia


Electrosurgery is the application of a high-frequency electric current to biological tissue as a means to cut, coagulate, desiccate, or fulgurate
Fulguration

Fulguration, also called electrofulguration, is a procedure to destroy tissue using an electric current....
 tissue. (These terms are used in specific ways for this methodology-- see below). Its benefits include the ability to make precise cuts with limited blood loss. Electrosurgical devices are frequently used during surgical operations helping to prevent blood loss in hospital operating rooms or in outpatient procedures.

In electrosurgical procedures, the tissue is heated by an electric current
Electric current

Electric current is the flow of electric charge. The electric charge may be either electrons or ions.The International System of Units unit of electric current intensity is the ampere....
. Although electrical devices may be used for the cauterization
Cauterization

The medical practice or technique of Cauterization is a medical term describing the burn of the body to remove or close off a part of itin a process called Cautery which destroys some tissue
 of tissue in some applications, electrosurgery is usually used to refer to a quite different method than electrocautery. The latter uses heat conduction
Heat conduction

Heat conduction or thermal conduction is the spontaneous heat transfer through matter, from a region of higher temperature to a region of lower temperature, and acts to equalize temperature differences....
 from a probe heated by a direct current (much in the manner of a soldering iron), whereas electrosurgery uses alternating current to directly heat the tissue itself.

Often electrosurgery is mistakenly referred to as diathermy. Unlike Ohmic heating by electric current passing through the conductive tissue in conventional electrosurgery, diathermy means dielectric heating, produced by rotation of molecular dipoles in high frequency alternating electric field. This effect is most widely used in microwave oven
Microwave oven

A microwave oven, or a microwave, is a kitchen appliance that cookings or heats food by dielectric heating. This is accomplished by using microwave radiation to heat water and other dipole within the food....
s which operate at GHz frequencies.

Electrosurgery is commonly used in dermatological, gynecological, cardiac, plastic, ocular, spine, ENT, orthopedic, urological, neuro- and general surgical procedures.

Electrosurgery is performed using an Electrosurgical Generator (also referred to as Power Supply or Waveform Generator) and a handpiece including one or several electrodes, sometimes referred to as an RF Knife.

History


Development of the first commercial electrosurgical device is credited to Dr. William T. Bovie, who worked on it from 1914 to 1927 while employed at Harvard University
Harvard University

Harvard University is a private university in Cambridge, Massachusetts, Massachusetts, United States, and a member of the Ivy League. Founded in 1636 by the colonial Massachusetts legislature, Harvard is the Colonial Colleges institution of higher learning in the United States....
  The first use of an electrosurgical generator in operating room occurred on October 1, 1926. The surgery was performed by Dr. Harvey Williams Cushing.

Tissue heating by electric current

When voltage is applied across the material it produces electric field
Electric field

In physics, the space surrounding an electric charge or in the presence of a time-varying magnetic field has a property called an electric field ....
 which exerts force on charged particles. A flow of free charge carriers – electrons and ions - is called electric current
Electric current

Electric current is the flow of electric charge. The electric charge may be either electrons or ions.The International System of Units unit of electric current intensity is the ampere....
. In metals and semiconductors the charge carriers are primarily electrons, whereas in liquids the charge is carried predominantly by ions. Electrical conduction in biological tissues is primarily due to the conductivity of the interstitial fluids, and thus is predominantly ionic. Transition between the electronic and ionic conduction is governed by electrochemical processes
Electrochemistry

Electrochemistry is a branch of chemistry that studies chemical reactions which take place in a solution at the interface of an electron Electrical conductor and an ionic conductor , and which involve electron transfer between the electrode and the electrolyte or species in solution....
 at the electrode-electrolyte interface. Value of electric current, I, is determined by the applied voltage, V, and material’s resistance, R, according to Ohm's law
Ohm's law

Ohm's law applies to electrical circuits; it states that the electric current through a conductor between two points is directly Proportionality to the potential difference or voltage across the two points, and inversely proportional to the Electrical resistance between them....
: Electric current of a constant polarity is referred to as Direct Current (DC). A current of alternating polarity is referred to as Alternating Current
Alternating current

In alternating current the movement of electric charge periodically reverses direction. An electric charge would for instance move forward, then backward, then forward, then backward, over and over again....
 (AC). Its frequency
Frequency

Frequency is the number of occurrences of a repeating event per unit time. It is also referred to as temporal frequency.The period is the duration of one cycle in a repeating event, so the period is the reciprocal of the frequency....
 is measured in cycles/second or Hertz (Hz).

Current flowing through a resistor causes the generation of Joule heating
Joule heating

Joule heating, also known as ohmic heating and resistive heating, is the process by which the passage of an electric current through a conductor releases heat....
. In other words, the resistance of the tissue converts the electric energy
Energy

In physics, energy is a scalar physical quantity that describes the amount of Work_ that can be performed by a force. Energy is an attribute of objects and systems that is subject to a conservation law....
 of the voltage source into heat (thermal energy) which causes the tissue temperature to rise. The deposited electric power
Electric power

Electric power is defined as the rate at which electrical energy is transferred by an electric circuit. The SI unit of power is the watt .When electric current flows in a circuit, it can transfer energy to do mechanical work or work ....
 (energy per time) can be calculated using: where P represents the electric power, typically measured in Watt
WATT

WATT is a radio station broadcasting a News radio-Talk radio-Sports radio format. Licensed to Cadillac, Michigan, it first began broadcasting in 1945....
s.

In absence of heat conduction
Heat conduction

Heat conduction or thermal conduction is the spontaneous heat transfer through matter, from a region of higher temperature to a region of lower temperature, and acts to equalize temperature differences....
, the rate of temperature
Temperature

In physics, temperature is a physical property of a Physical system that underlies the common notions of hot and cold; something that feels hotter generally has the greater temperature....
 rise, dT/dt, in a heated object is proportional to the deposited power P, and inversely proportional to its heat capacity, which is in turn proportional to the mass
Mass

In physical science, mass refers to the degree of acceleration a body acquires when subject to a force: bodies with greater mass are accelerated less by the same force....
 m of the object and its specific heat capacity
Specific heat capacity

Specific heat capacity, also known simply as specific heat, is the measure of the energy required to increase the temperature of a of a substance by a certain Celsius#Temperatures_and_intervals....
 c:


Larger amount of heat is required to increase the temperature of a heavier object. Thus when heat is generated in a small region of an object, the temperature of that localized region will rise much faster than if the same amount of heat is evenly dispersed over the entire object.

Current density
Electric current

Electric current is the flow of electric charge. The electric charge may be either electrons or ions.The International System of Units unit of electric current intensity is the ampere....
, j is a measure of the concentration of electric current. A higher current density results in a higher concentration of Joule heating
Joule heating

Joule heating, also known as ohmic heating and resistive heating, is the process by which the passage of an electric current through a conductor releases heat....
. Power density generated by electric current in the material, p is proportional to the square of the current density, and to the material's resistivity
Resistivity

Electrical resistivity is a measure of how strongly a material opposes the flow of electric current. A low resistivity indicates a material that readily allows the movement of electrical charge....
, g:

In absence of heat conduction
Heat conduction

Heat conduction or thermal conduction is the spontaneous heat transfer through matter, from a region of higher temperature to a region of lower temperature, and acts to equalize temperature differences....
, the rate of local temperature rise is proportional to the power density, p, produced in that region of tissue, and inversely proportional to its specific heat capacity
Specific heat capacity

Specific heat capacity, also known simply as specific heat, is the measure of the energy required to increase the temperature of a of a substance by a certain Celsius#Temperatures_and_intervals....
 and density
Density

The density of a material is defined as its mass per unit volume. The symbol of density is ....
  .


Electrical stimulation of neural and muscle cells

Neural and muscle
MUSCLE

MUSCLE is public domain, multiple sequence alignment software for protein and nucleotide sequences.MUSCLE is integrated into UGENE bioinformatics tool as a plugin....
 cells are electrically-excitable, i.e. they can be stimulated by electric current. In human patients such stimulation may cause acute pain, muscle spasms, and even cardiac arrest
Cardiac arrest

A cardiac arrest, also known as cardiopulmonary arrest or circulatory arrest, is the abrupt cessation of normal circulation of the blood due to failure of the heart to contract effectively during Systole ....
. Sensitivity of the nerve and muscle cells to electric field is due to the voltage-gated ion channel
Voltage-gated ion channel

Voltage-gated ion channels are a class of Transmembrane protein ion channels that are activated by changes in electrical potential difference near the channel; these types of ion channels are especially critical in neurons, but are common in many types of cell ....
s present in their cell membrane
Cell membrane

The cell membrane is the interface between the cellular machinery inside the cell and the fluid outside.It is a semipermeable lipid bilayer found in all cell ....
s. Stimulation threshold does not vary much at low frequencies (so called rheobase
Rheobase

In neuroscience, rheobase is the minimal electric current of infinite duration that results in an action potential or the contraction of a muscle....
 - constant level). However, the threshold starts increasing with decreasing duration of a pulse (or a cycle) when it drops below a characteristic minimum (so called chronaxie
Chronaxie

In the mathematical description of the functioning of the nervous system, the chronaxie is the minimum time over which an electric current double the strength of the rheobase needs to be applied, in order to stimulate a muscle fiber or nerve cell....
). Typically, chronaxie
Chronaxie

In the mathematical description of the functioning of the nervous system, the chronaxie is the minimum time over which an electric current double the strength of the rheobase needs to be applied, in order to stimulate a muscle fiber or nerve cell....
 of neural cells is in the range of 0.1 - 10 ms, so the sensitivity to electrical stimulation (inverse of the stimulation threshold) decreases with increasing frequency in the kHz range and above. (Note that frequency of the alternating electric current is an inverse of the duration of a single cycle). To minimize the effects of muscle and neural stimulation, electrosurgical equipment typically operates in the radio frequency
Radio frequency

Radio frequency is a frequency or rate of oscillation within the range of about 3 Hz to 300 GHz. This range corresponds to frequency of alternating current electrical signals used to produce and detect radio waves....
 (RF) range of 100 kHz to 5 MHz.

Operation at higher frequencies also helps minimizing the amount of hydrogen and oxygen generated by electrolysis of water
Electrolysis of water

Electrolysis of water is the decomposition of water into oxygen and hydrogen gas due to an electricity being passed through the water. This electrolysis is used in some industrial applications when hydrogen is needed....
. This is especially important consideration for applications in liquid medium in closed compartments, where generation of gas bubbles may interfere with the procedure. For example, bubbles produced during an operation inside an eye may obscure a field of view.

Common electrode configurations

There are several commonly used electrode configurations or circuit topologies:

In bipolar configuration the voltage is applied to the patient using a pair of similarly-sized electrodes. For example, special forceps
Forceps

Forceps are a handheld, hinged instrument used for grasping and holding objects. Forceps are used when fingers are too large to grasp small objects or when many objects need to be held at one time while the hands are used to perform a task....
, with one tine connected to one pole of the AC generator and the other tine connected to the other pole of the generator. When a piece of tissue is held by the forceps, a high frequency electric current flows from one to the other forceps tine, heating the intervening tissue.

In monopolar configuration the patient lies on top of the return electrode, a relatively large metal plate or a flexible metalized plastic pad which is connected to the return electrode of the AC source. The surgeon uses a pointed probe
Probe

Probe can mean the following things:*Space probe, a scientific space exploration mission.*An Anal probe, used medically.*Langmuir probe, an electrode inserted into a plasma to determine temperature and density....
 to make contact with the tissue. The electric current flows from the probe tip, through the body to the return electrode, and then back to the electrosurgical generator. Since electric current spreads from the pointed electrode as it enters the body the current density
Electric current

Electric current is the flow of electric charge. The electric charge may be either electrons or ions.The International System of Units unit of electric current intensity is the ampere....
 is rapidly (quadratically) decreasing with distance from the electrode. Since the rate of heating is proportional to the square of current density, the heating occurs in a very localized region, only near the probe tip. On an extremity such as a finger, there is limited cross-sectional area for the return current to spread across, which might result in higher current density and some heating throughout the volume of the extremity.

There is also a common intermediate configuration, when both electrodes are located on the same probe, but the return electrode is much larger than the active one. Since current density is higher in front of the smaller electrode, the heating and associated tissue effects take place only (or primarily) in front of the active electrode, and exact position of the return electrode on tissue is not critical. Sometimes such configuration is called sesquipolar, even though the origin of this term in Latin (sesqui) means a ratio of 1.5 .

Relatively low-powered high frequency electrosurgery can be performed on conscious outpatients with no return electrode at all . Operating with no return electrode is possible, because at the very high frequencies and low currents, the self-capacitance
Capacitance

In electromagnetism and electronics, capacitance is the ability of a body to hold an electrical charge.Capacitance is also a measure of the amount of electric charge stored for a given electric potential....
 of the patient's body (which is between the patient's body and the machine's return potential) is large enough to allow the resulting displacement current
Displacement current

In electromagnetism, displacement current is a quantity that is defined in terms of the rate of change of electric displacement field. Displacement current has the units of electric current and it has an associated magnetic field....
 to act as a return path. One example of such a machine is called a hyfrecator
Hyfrecator

A hyfrecator is a medical apparatus used in electrosurgery. It works by emitting low-power high-frequency A.C. electrical pulses via a probe, directly to the affected area of the body....
.

An accidental additional return path through an earth-ground provides a danger of a burn at a site far away from the probe electrode, and for this reason single-electrode devices are used only on conscious patients who would be aware of such complications, and only on carefully insulated tables.

Electrosurgical modalities


In Cutting mode electrode touches the tissue, and sufficiently high power density is applied to vaporize its water content. Since water vapor is not conductive under normal cirumstances, electric current cannot flow through the vapor layer. Energy delivery beyond the vaporization threshold can continue if sufficiently high voltage is applied (> +/-200 V) to ionize vapor and convert it into a conductive plasma. Vapor and fragments of the overheated tissue are ejected, forming a crater .

Coagulation is performed using waveforms with lower power, generating heat insufficient for explosive vaporization, but producing a thermal coagulum instead.

Electrosurgical desiccation
Desiccation

Desiccation is the state of extreme dryness, or the process of extreme drying. A desiccant is a hygroscopic substance that induces or sustains such a state in its local vicinity in a moderately-well sealed container....
 occurs when the electrode touches the tissue open to air, and the amount of generated heat is lower than that required for cutting. The tissue surface dries out and forms a coagulum.

In fulguration
Fulguration

Fulguration, also called electrofulguration, is a procedure to destroy tissue using an electric current....
 mode the electrode is held away from the tissue, so that when the air gap between the electrode and the tissue is ionized, an electric arc
Electric arc

An electric arc is an electrical breakdown of a gas which produces an ongoing Plasma Electrostatic discharge, resulting from a current flowing through normally Electrical conductance media such as air....
 discharge develops. In this approach the burning to the tissue is more superficial, because the current is spread over the tissue area larger than the tip of electrode. Under these conditions, superficial skin charring or carbonization is seen over a wider area than when operating in contact with the probe. Ionization of an air gap requires voltage in the kV range.

Besides the thermal effects in tissue, electric field can produce pores in the cellular membranes - a phenomenon called electroporation
Electroporation

Electroporation, or electropermeabilization, is a significant increase in the electrical conductivity and permeability of the cell membrane caused by an externally applied electrical field....
. This effect may affect cells beyond the range of thermal damage.

Electrosurgical waveforms

Different waveform
Waveform

Waveform means the shape and form of a signal such as a wave moving in a solid, liquid or gaseous medium.In many cases the medium in which the wave is being propagated does not permit a direct visual image of the form....
s can be used for different electrosurgical procedures. For cutting, a continuous single frequency sine wave
Sine wave

The sine wave or sinusoid is a function that occurs often in mathematics, physics, signal processing, hearing , electrical engineering, and many other fields....
 is often employed. Rapid tissue heating leads to explosive vaporization of interstitial fluid
Interstitial fluid

Interstitial fluid is a solution which bathes and surrounds the cells of multicellular animals. It is the main component of the extracellular fluid, which also includes Blood plasma and transcellular fluid....
. If the voltage is sufficiently high (> 400 V peak-to-peak) the vapor sheath is ionized, forming conductive plasma
Plasma (physics)

In physics and chemistry, plasma is a partially ionized gas, in which a certain proportion of electrons are free rather than being bound to an atom or molecule....
. Electric current continues to flow from the metal electrode through the ionized gas into the tissue. Rapid overheating of tissue results in its vaporization, fragmentation and ejection of fragments, allowing for tissue cutting. In applications of a continuous wave the heat diffusion typically leads to formation of a significant thermal damage zone at the edges of the lesion. Open circuit voltage in electrosurgical waveforms is typically in the range of 300 - 10,000 V peak-to-peak.

Higher precision can be achieved with pulsed waveforms . Using bursts of several tens of microseconds in duration the tissue can be cut, while the size of the heat diffusion zone does not exceed the cellular scale. Heat accumulation during repetitive application of bursts can also be avoided if sufficient delay is provided between the bursts, allowing the tissue to cool down . The proportion of ON time to OFF time can be varied to allow control of the heating rate. A related parameter, duty cycle
Duty cycle

In telecommunications and electronics, the duty cycle is the fraction of time that a system is in an "active" state. In particular, it is used in the following contexts:...
, is defined as the ratio of the ON time to the period (the time of a single ON-OFF cycle). In the terminology of electrical engineering
Electrical engineering

Electrical engineering, sometimes referred to as electrical and electronic engineering, is a field of engineering that deals with the study and application of electricity, electronics and electromagnetism....
, this process of altering an amplitude of a periodic waveform is called modulation
Modulation

In telecommunications, modulation is the process of varying a Periodic function waveform, i.e. a tone, in order to use that signal to convey a message, in a similar fashion as a musician may modulate the tone from a musical instrument by varying its volume, timing and Pitch ....
.

For coagulation, the average power is typically reduced below the threshold of cutting. Typically, sine wave is turned ON and OFF in a rapid succession. The overall effect is a slower heating process, which causes tissue to coagulate.

Many modern electrosurgical generators provide sophisticated waveforms with power adjusted in real time, based on changes of the tissue impedance.

Prevention of unintended burns in patients

For high power surgical uses during anesthesia the monopolar modality relies on a good electrical contact between a large area of the body (typically at least the entire back of the patient) and the return electrode. If contact with the return pad is insufficient, severe burns (3rd degree) can occur in areas of poor contact with the return pad, or with metal objects in contact with Earth-ground serving as an unintended (capacitative) return path.

To prevent unintended burns, the skin is cleaned and a conductive gel is used to enhance contact. Proper electrical grounding practices must be followed in the electrical wiring of the building. It is also recommended to use a newer electrosurgical unit that includes alarms for ground circuit interruption. Grounding pads should always have full contact with the skin and be placed on the same side of the body and close to the body part where the procedure is occurring.

If there is any metal in the body of the patient, the grounding pad is placed on the opposite side of the body from the metal and be placed between the metal and the operation site. This precents current from passing selectively through metal on the way to ground. For example, for a patient who has had a right sided hip replacement who is scheduled for surgery, the grounding pad is placed on the left side of the body on the lateral side of the lower abdomen, which places the grounding pad between the location of the metal and the surgical site and on the opposite side from the metal. If there is metal on both sides of the body, the grounding pad is placed between the metal and the procedure site when possible. Common grounding pad locations include lateral portions of the outer thighs, abdomen, back, or shoulder blades.

The use of the bipolar option does not require the placement of a grounding pad because the electrical current does not flow through the bulk of the body.

Electrosurgery should only be performed by a physician who has received specific training in this field and who is familiar with the techniques used to prevent burns.

See also

  • Cryosurgery
    Cryosurgery

    Cryosurgery is the application of extreme cold to destroy abnormal or diseased tissue. The term comes from the Greek words cryo and surgery meaning "hand work" or "handiwork"....
  • Laser surgery
    Laser surgery

    Laser surgery is surgery using a laser to cut Tissue instead of a scalpel. Examples include the use of a laser scalpel in otherwise conventional surgery, and soft tissue laser surgery, in which the laser beam vaporizes soft tissue with high water content....
  • Electrocautery
    Cauterization

    The medical practice or technique of Cauterization is a medical term describing the burn of the body to remove or close off a part of itin a process called Cautery which destroys some tissue
  • Diathermy
  • microwave minimaze procedure


Manufacturers of Electrosurgical Equipment


In alphabetical order.
  • , a Bovie Company, manufacturer of the Aaron series of devices.
  • , manufacturer of the adeor RF4 Radiofrequency Surgical Unit.
  • , a division of B. Braun, manufacturer of the BipoJet Bipolar Product Line, as well as other mono and bipolar devices.
  • , manufacturer of the Surgi-Max 4.0 MHz Radiofrequency Surgical Unit
  • , manufacturer of the Surgitron Dual Frequency 4.0 MHz device.
  • , manufacturer of the VIO series of devices.
  • , manufacturer of the Plasmacision and Plasmaknife devices.
  • , manufacturer of the MEGA 2000 of device.
  • , Manufacturer of the ElectroSurgical Forceps.
  • ,Manufacturer of the PEAK Surgical System: Pulsar generator and PlasmaBlade handprobes.
  • , manufacturer of the Conmed Hyfrecator 2000 and the Meyer-Haake RadioSurg devices.
  • , manufacturer of the SS-601MCa and many other devices.
  • , a division of Tyco Healthcare Group LP, manufacturer of the Force FX and Force EZ devices.


External links

In alphabetical order
  • , Richard J Motley, Schuco International Ltd. a primer for low-powered outpatient dermatological devices, such as the Hyfrecator 2000 device.
  • , Barry L. Hainer M.D., Richard B. Usatine, M.D., American Family Physician (Journal of the American Academy of Family Physicians), 2002 Oct 1;66(7):1259-66.
  • Online Journal of he Biomedical Engineering Association of Ireland (BEAI), May 1997.
  • Kristin Cermak, Erica Gorbutt, Brian Schielke, Trey Souchock * at The Laser Training Institute of the Professional Medical Education Association, Inc, Ohio.
  • at Valleylab, a manufacturer of electrosurgical equipment.
  • , Judith Lee, Contributing Editor, Outpatient Surgery Magazine, February, 2002.
  • , Electro Surgical Instruments Supplies