We have a Leeman Labs PS 950 ICP AES using PSWin. Since after crashing the program due to DOS storage overload the instrument has not worked. The machine responds to the program prompts, but the results are not correct. When running a Hg full peak routine the results show flat (=zero) plots. When scanning any element at any concentration the emission counts are the same, at blank values (5 for Ti and 191 for P), and plots show no peaks. Nothing was changed in the instrument from the day when it worked perfectly well to the next morning when the program crashed. Any suggestions as to possible solutions or even diagnosis?
Save
Cancel
Reply
replied to:  rhb1
Vito59
Replied to:  We have a Leeman Labs PS 950 ICP AES using PSWin....
Please contact me directly, I have some suggestions but need more information. My e-mail is VMackiewicz@teledyne.com
Save
Cancel
Reply
replied to:  Vito59
rhb1
Replied to:  Please contact me directly, I have some suggestions but need more...
Mr. Mackiewicz, thank you for your offer to help. You have already helped me and your suggestion did solve our problem! We are up and running again - the problem was the interlock, as you had indicated. Thank you again.
Save
Cancel
Reply
replied to:  rhb1
Vito59
Replied to:  Mr. Mackiewicz, thank you for your offer to help. You have...
I thought the problem sounding familiar but I wanted to make sure it wasn't someone else